Hottop probe probes reading difference
Robert Mayrand
I just did the upgrade on my hottop, and strangely if i run the roaster empty, the two probes wich are less than two inches apart show a 50f difference!!!!. I would have understand with the roaster loaded, but empty should?nt they show about the same temp or just a few degree difference because of their location.

Rob
 
renatoa
Maybe they are different types, K and J, and software isn't set accordingly.
The offset value suggest this behaviour, the voltage output a J-type TC at 150 is very very close to K-type TC at 200, in the 8 mV ballpark.
 
Robert Mayrand
Hottop specification mention that this is a dual k thermocouple model so that should?nt be the problem.
 